Dear Users
As many of you will know, the MEG scanner underwent preventative
maintenance earlier this month.
We are happy to report that this work was carried out successfully. Due
to the re-cooling of the sensors and some calibration work the scanner
is working well. The noise floor has improved considerably.
However, one of the inherent dangers of this thermal cycling is the
potential loss of sensors. Until now we had been very fortunate in that
we had not lost any channels but this is no longer the case.
Users should take note of the following:
1) "Dead" channels
There are 3 channels in the MEG scanner that do not work. These channels
are A55, A84 and A237. There had recently been suspicion about the
quality of the data from channel 237 and during the maintenance it was
confirmed that this channel is not working correctly. On a more positive
note this is an electronics problem and will be solved once our new MEG
electronics are in place. For now, anyone collecting new MEG data should
exclude channels A55, A84 and A237 from their analyses. There will be
some minor changes to some YNiC software to allow for channel exclusions.
2) REI board error (ONLY for MEG operators)
Although currently stable, we are aware of a problem with one of the
electronics boards (the REI board). MEG operators will know when the REI
board is not working because of an error message that appears in the ACP
window on the Sun work station. For this reason the ACP, DCP, AVP and
VIP windows must be open when you are acquiring MEG data so that any
problems with the MEG scanner can be tracked. If any of these windows
are not open, they can be opened from the "acquisition setup" window in
the database on the Sun workstation. Click on the "utilities" menu and
from here open the relevant window/s.
Please record any faults in the "MEG error log book" (left-hand side of
the Sun machine). In the event of an REI board error, please contact a
member of YNiC staff.
Please do not hesitate to contact me if you have any further questions
about these matters.
Best wishes
Rebecca
--
************************************************************************